I am using an OPT301 combination photodiode transimpedance amp to pick up the presence of a fluorescent chemical in a small (<40um) fluid filled channel. The light level from this tiny amount of the chemical is very small and consequently I have been using the OPT301 with an external 10G resistor for gain. This does work although I'd like to push it maybe 10x further while minimizing the noise. Here in lies my problem or, more to the point, my several problems:
1. 10G is as large a surface mount resistor as I can easily obtain. I was planning therefore to use a T-Network to accomplish a higher overall gain with smaller value resistors. I have read however that this will increase my 1/f noise more than a single resistor would and was hoping for any advice the forum might have in limiting this effect.
2. As I have increased the gain on the OPT301, I have seen an increasingly long settling time in response to the signal. My signal of interest is <10Hz as no more than 10 bands of the fluorescent chemical will be able to pass the detector in under a second. The rise time of the detector appears more than adequate for this rate. However, once a signal is detected, the drop off back to zero can take a considerable amount of time. Am I correct in thinking this is a direct result of the extreme gain? What is the best way to eliminate this effect?
Thanks,
Nate Oborny