Hi there,
I am doing some tests on the LMP91000EVM. I found that the reaction time of using the 91000 AFE is much slower than using a discrete circuit. I have tried on both CityTech sensors and Alphasense, both are the same. the T90 is more than 80 sec., much longer than the regulation requirement, which is <25 sec. I have already taken off the reading averaging options, and the reading was the raw voltage. Is there anything I could have been doing wrong?
Thanks,
Jian