Other Parts Discussed in Thread: INA116, LMP91200
Hello,
I am a chemistry student, and I'm currently trying to build a custom pH meter. I've decided to use the INA116 amp because it provides its own guards, and has an increddibly low input bias current. This is perfect for pH electrodes that have internal resistance of several hundred megaohms. However after reading the datasheet I have a few questions about this IC.
A) First of all I'd like to know a few details about the input bias current path. The datasheet suggests (page 8, figure 3) that some resistors be connected between each amplifier input pin, and the ground. According to that schematics I see two 100 MOhm resistors connected to each input, and the ground. Is this correct - are these really needed in my case? The sensor I will be using is a pH electrode, and they have internal resistances up to 1000 MOhm. So basically what we just did here is connected a 200 MOhm resistor in parallel to the INA116 inputs.
This means we just added a low-resistance bypass path for the already-weak current that is originationg from the electrode. We are basically measuring the voltage drop on the 200 MOhm resistor, not the pH electrode. The resistance/impedance of the INA is so high that it can be neglected. The pH electrode gives a voltage of 413.7 mV at best, and let's assume the internal resistance is 1000 MOhm. The total resistance of the current path is therefore 1000 + 200 = 1200 MOhm. According to ohm's law (I=U/R), the current at this voltage will be 3.4475E-10 A. That's about 0.3 nA or 344.75 pA. The voltage drop will therefore be U=I*R, and the resistor causes the voltage drop of 68.95 mV, where the electrode has a voltage drop of 344.75 mV due to its internal resistance.
We're therefore measuring 69 mV instead of 413 mV - that's about 83% rel. error we are causing by adding these two resistors here!
What os the proper solution to this?
B) The datasheet also states that guards should fully encircle both input pins on both sides of the PCB. In this case how are the former resistors (assuming we're measuring a lower-impedance source, so that the previous question doesn't apply) supposed to be connected to the input pins? Can they be connected over the guard?
C) What is the maximum output current that INA116 is able to provide? According to the datasheet on pg 2, this would be +5 / -12 mA. Is this correct? If so, what is the proper way to level-shift the voltage to adapt it for a 5V ADC like AD7791? A resistor voltage divider can cause a large error in the ADC measurement. Too low resistor values cause a large voltage drop. Too high, and the ADC doesn't get enough current. What's the proper way to voltage-shift, if the INA is being powered by a single-end power supply of 12V, and has a virtual ground of 6.0V? The ADC is on 5VDC (12V => LM7805 => 5V).
D) Can you offer any other tips&tricks, hints, comments or suggestions about pH meter design?
Thanks a lot.
Best regards!