PGA309: Offset

Part Number: PGA309

Hello,

I use your PGA309 in a prototype I made. The component seems to operate correctly. I compensate our pressure sensor in pressure and temperature, but  sometimes there is an output offset after a 125°C temperature step (-20mV). Our pressure sensors are used to provide a 0 to 50mV FS and I tune the PGA to reach a 1V to 5V output. I tune the output with ±2mV accuracy from -40 to 125°C in Test Mode, and then I flash the EEPROM. After that I calibrate all the sensor in stand-alone (pressure and temperature) and the result seems correct but with a -20mV offset. This does not come from our pressure cell ...

have you ever seen this problem ?*

Thanks

Guillaume

  • Hi Guillaume,

    The input referred offset of the PGA309 is 50uV with a typical drift of +/-2uV/C. Over a temperature change from 25C to 125C, the typical input referred offset error change should be in the range of ~200uV.

    What are you gain Settings for the Front End PGA , the Output Amp Gain settings? What is the coarse offset DAC setting and Reference settings? Do you have an schematic of the prototype?

    If you can detail the steps you have followed during the calibration process, this may be helpful

    Most issues we encounter are due to issues related to the calibration sequence and/or software. If everything works properly and the error only occurs once in a while, a noise disturbance or glitch could be occurring during the calibration process.

    Thank you,

    Best Regards,

    Luis
  • Hello, 

    I can't send our schematic on this forum, but these are the register values used and the EEPROM values.

    For the moment there is no program. I compensate manualy (using sboc320c.zip to change the registers manualy and to program the EEPROM)

    The steps I do :

    1/ Program EEPROM with standard values (coarse offset, Input Gain, Out put Gain, Temperature mode....)

    2/ PGA Test Mode

    3/ Temperature @-55, tune Gain DAC and Offste DAC (pressuse : 0 and FS) and record temperature Hex value, Gain DAC and Offste DAC

    4/ Temperature @-10, tune Gain DAC and Offste DAC (pressuse : 0 and FS) and record Temperature Hex value, Gain DAC and Offste DAC

    5/ Temperature @35, tune Gain DAC and Offste DAC (pressuse : 0 and FS) and record Temperature Hex value, Gain DAC and Offste DAC

    6/ Temperature @80, tune Gain DAC and Offste DAC (pressuse : 0 and FS) and record Temperature Hex value, Gain DAC and Offste DAC

    7/ Temperature @125, tune Gain DAC and Offste DAC (pressuse : 0 and FS) and record Temperature Hex value, Gain DAC and Offste DAC

    8/ Calculate with exel a polynome for the Offset DAC vs Temperature and a polynome for the Gain DAC vs Temperature in order to calculate intermediate gain and offset values between each tuned temperature (to use all temperature steps of the EEPROM table.

    9/ Calculate with exel all the values to send to the EEPROM

    10/ PGA Stand-Alone Mode 

    11/ Program the EEPROM

    12/ Measure the sensor 

    When I tune the sensor in test mode, the error is less than 1mV for each temperature or pressure step. But when  I measure the sensor a few days after the result is 10mV lower ...

    REGISTRE 0 5449 21577 TI Flag
    REGISTRE 1 0000 0 ZeroDAC
    0
    REGISTRE 2 0000 0 GainDAC
    0,3333333
    REGISTRE 3 0700 1792 Lin Range (EXS)
    Range 0
    Vexc Enable (EXEN)
    Enable
    Internal Vref (RS)
    2,5
    Internal Vref (REN)
    Internal Ref
    Bv (%FS)
     
    REGISTRE 4 0613 1555 OWD
    Enable
    Gain Out GO
    2
    Front-End Gain GI
    64
    Coarse Offset
    -0,006375
    REGISTRE 5 0000 0 CLK mode (Noise)
    00
    EXTEN (Enable External faul comparator)
    Disable
    INTEN (Enable Internal faul comparator)
    Disable
    EXTPOL (Out polarity when external fault)
    Force Low
    INTPOL (Out polarity when internal fault)
    Force Low
    OUEN (Over/Under scale Limite)
    Disable
    Overscale Threshold
    2,427
    Overscale Threshold
    0,064
    REGISTRE 6 140D 5133 ADC2X (Temp conversion speed EXTERNAL sense)
    6ms
    ADCS (Temp conversion START conversion)
    Single Start
    ISEN (TEMPin current source 7µA)
    Disable
    CEN (Temp Continous Conversion Mode)
    Enable
    TEN (Intenal Temp Mode)
    Disable
    AREN (Intenal Temp Reference)
    Disable
    RV (External Temp Reference select)
    VREF
    M (Temp pin selection)
    TEMPin,GNDa
    G (Temp Gain)
    8V/V
    R (Temp Resolution)
    Low

    register.xlsx

    Addr0=21577
    Addr2=0
    Addr4=0
    Addr6=1792
    Addr8=1555
    Addr10=0
    Addr12=5133
    Addr14=35478
    Addr16=3649
    Addr18=15548
    Addr20=61989
    Addr22=3760
    Addr24=355
    Addr26=59814
    Addr28=3854
    Addr30=338
    Addr32=60027
    Addr34=3949
    Addr36=321
    Addr38=60227
    Addr40=4043
    Addr42=305
    Addr44=60424
    Addr46=4138
    Addr48=291
    Addr50=60623
    Addr52=4232
    Addr54=275
    Addr56=60819
    Addr58=4327
    Addr60=261
    Addr62=61020
    Addr64=4422
    Addr66=243
    Addr68=61216
    Addr70=4516
    Addr72=229
    Addr74=61415
    Addr76=4611
    Addr78=213
    Addr80=61615
    Addr82=4705
    Addr84=199
    Addr86=61813
    Addr88=4800
    Addr90=181
    Addr92=62009
    Addr94=4894
    Addr96=169
    Addr98=62208
    Addr100=4989
    Addr102=151
    Addr104=62407
    Addr106=5083
    Addr108=133
    Addr110=62606
    Addr112=5194
    Addr114=120
    Addr116=62819
    Addr118=32767
    Addr120=0
    Addr122=9331

  • Hi Guillaume,

      When compensating for the offsets and gain errors manually, which algorithm are you using to perform the Gain DAC and offset DAC corrections? Is it the same as the algorithm attached below?  The document attached below details the mathematics and register writes behind the calibration process on the PGA309EVM. 

    PGA309_calibration_procedure.pdf

    What type of meter are you using to perform the measurements during the calibration process?

    Please let me know if you wish that I contact you directly via email so we can review your schematic.  I will also continue reviewing the register settings provided.

    Thank you and Regards,

    Luis

  • Hi,

    The algorithm I use is quite the same your document. The only difference is that I calibrate at five temperature steps and I use GainDAC and OffsetDAC polynomes to add values in the table-lockup (EEPROM).

    I use a keithley 2750 with a switch card (it s like a 6 digits +1/2 DMM)

    yes, I would like to be contacted by email.

    Did you find something abnormal in the registers I provided ?

    Thanks

    Guillaume

  • Hi Guillaume,

    I will contact you directly via email to per your request.

    Thanks and Regards,

    Luis
  • HI Guillaume,

    I have contacted via email; but have not heard from you... Please feel free to contact me via email and/or post back in here if you have additional queries.

    Thank you,

    Luis