I am using the INA169 to monitor current flow in a test & measurement application. When, for example, 10mV is read across the shunt resister, and the gain is set for x100, then I should expect the device to produce a 1V output. Is this correct?
This thread has been locked.
If you have a related question, please click the "Ask a related question" button in the top right corner. The newly created question will be automatically linked to this question.
I am using the INA169 to monitor current flow in a test & measurement application. When, for example, 10mV is read across the shunt resister, and the gain is set for x100, then I should expect the device to produce a 1V output. Is this correct?
Two Few,
Ideally, you are correct. Since 10mV is a relatively low voltage, offset errors of the INA169 will contribute to the offset at the output. The 1mV input offset voltage (max rating) would potentially contribute a 100mV offset at the output. So just looking at the offset errors, a 10mV input could give you a 0.9V to 1.1V output. It would be more correct to say that a 10mV change in the shunt voltage would produce a 1V change in the output voltage (now ignoring the possible gain errors). To minimize the effect of offset errors, use a larger full-scale shunt voltage.
Regards, Bruce.