LMX2492: DDS-PLL Phase Noise Measurement VS Simulation

Part Number: LMX2492

I’ve been measuring the phase noise of the signal generated with the following setup using an Agilent E4407B spectrum analyser with Phase Noise measurement mode:

R&S SMU200A vector signal generator used as 400 MHz sampling clock signal for ...

ADI AD9956 DDS which is used to provide 100 MHz reference signal to ...

TI_LMX2492EVM (TI LMX2492 PLL + Macom MAOC-009265) generating a 9.5625 GHz signal which is input to ...

ADI HMC814LC3B x2 multiplier on a custom board to give 19.125 GHz output signal

 

I have also simulated the phase noise level in TI PLLatinum, the table below compares the measured and simulated results:

Frequency Offset from Carrier (Hz)

Measured Phase Noise Level (dBc/Hz)

Simulated Phase Noise Level (dBc/Hz)

Disparity Between Measured/Simulated Phase Noise Levels (dB)

1e3

-65.8

-75.2

-9.2

1e4

-82.0

-84.5

-2.5

1e5

-86.2

-91.6

-5.4

1e6

-92.1

-94.3

-2.2

 

With the exception of the 1 KHz offset I got fairly good agreement, do you know why I’m seeing such a difference at 1 KHz?

For the simulation I took the R&S SMU200A phase noise level from its datasheet (https://cdn.rohde-schwarz.com/pws/dl_downloads/dl_common_library/dl_brochures_and_datasheets/pdf_1/SMU_dat-sw-en.pdf, page 13) and estimated DDS output phase noise as shown below:

Offset from Carrier Frequency (Hz)

Residual Phase Noise of DDS [at 100 MHz, estimated from datasheet] (dBc/Hz)

Phase Noise of Reference at 400 MHz (dBc/Hz)
[SMU200A]

Phase Noise of Reference at 100 MHz (dBc/Hz)

Total Phase Noise (dBc/Hz)

1000

-123

-114

-126.04

-121.25

1.00E+04

-131

-133

-145.04

-130.83

1.00E+05

-138

-136

-148.04

-137.59

1.00E+06

-142

-150

-162.04

-141.96

 

 

I took the phase noise of the Macom MAOC-009265 VCO from its datasheet, note that I didn’t change the loop filter from its default values. Also, I set up the simulation for 9.5625 GHz output and then added 6 dB to give simulated phase noise levels at 19.125 GHz. I could send through the TI PLLatinum simulation file.

Note that when I measured the phase noise of the SMU200A vector signal generator output at 400 MHz or DDS output at 100 MHz I didn’t get very sensible results so used results from datasheet/calculation to input these phase noise values into TI PLLatinum, I thought that maybe the phase noise of the Agilent E4407B spectrum analyser was impacting the measurement maybe:

  • Measured phase noise of SMU200A vector signal generator: 1 KHz = -81.01 dBc/Hz / 10 KHz = -100.55 dBc/Hz / 100 KHz = -123.80 dBc/Hz / 1 MHz = -126.80 dBc/Hz
  • Measured phase noise of 100 MHz output from DDS – 1 KHz = -82.16 dBc/Hz / 10 KHz = -101.03 dBc/Hz / 100 KHz = -122.76 dBc/Hz / 1 MHz = -127.30 dBc/Hz

When I measured the phase noise of the TI LMX2492EVM 4.8 GHz output with on board XO used as a reference – 1 KHz = -81.5 dBc/Hz / 10 KHz = -97.7 dBc/Hz / 100 KHz = -107.9 dBc/Hz / 1 MHz = -118.5 dBc/Hz.

Relative to the plot on page 7 of the TI LMX2492EVM user guide (copied below) the phase noise level is similar at 10 KHz/100 KHz/1 MHz offsets but is almost 10 dB higher at 1 KHz, i.e. similar difference to that observed with DDS-PLL measurement above – could it be that the Agilent E4407B isn’t measuring phase noise accurately at 1 KHz offset maybe?

  • Hi Peter,
    I think the difference comes from the spectrum purity of E4407B. We used E5052B to take the measurement. The phase noise of E5052B is 30dB better than the E-series spectrum analyzer.
  • Hi Noel,

    I think that could be the issue - the phase noise level of the E4407B is shown here (literature.cdn.keysight.com/.../5989-9815EN.pdf, page 8) for a centre frequency of 1 GHz. Although it isn’t stated could assume that at 1 KHz offset from carrier may be -78 dBc/Hz or so.
    Can see from my measurements above that there is a level of -81 dBc/Hz at 1 KHz offset for each of the below measurements: (i) SMU200A vector signal generator at 400 MHz, (ii) DDS at 100 MHz, (iii) TI LMX2492EVM at 4.8 GHz. The E4407B phase noise level would explain these measurements.

    For centre frequency of 19.125 GHz the phase noise level of E4407B isn’t stated, if this is the cause of the 9.2 dB disparity between simulated/measured results for full DDS-PLL chain measurement at 1 KHz offset then must be significantly higher than for a centre frequency of < 5 GHz, which could be the case