Dear Specialists,
My customer is considering ADS125H01 and has a question about relationship between external clock and sampling rate.
I would be grateful if you could advise.
---
I want to input a 6.144MHz external clock signal to the ADS125H01 and output 2400sps.
Is it possible to achieve this by setting bits 7:3 DR[4:0]DR of the Mode 0 (MODE0) Register (address = 02h) to 01010?
Is this understanding correct?
Or is it necessary to set the external clock frequency to 7.328MHz for outputting 2400SPS?
ーーー
I appreciate your great help in advance.
Best regards,
Shinichi
Hi Shinichi Inoue,
Or is it necessary to set the external clock frequency to 7.328MHz for outputting 2400SPS?
Yes this is correct. The data rates will scale with the clock frequency. If you provide a 6.144 MHZ clock, then the data rate will scale by a factor of 6.144 / 7.3728 = 0.83333
Therefore, the 2400 SPS data rate will actually be 2000 SPS
-Bryan
HI Bryan,
Thank you for your reply.
I understand when 6.144MHz CLK and set 2400SPS, it outputs 2000SPS.
Because 2400SPS is internal clock (7.3728MHz), it must be scaled by 6.144/7.3728.
I'll share the information with the customer.
I appreciate your great help and cooperation.
Best regards,
Shinichi