Hi Albert,
The VBIAS can be used is cases where you have a floating sensor voltage. VBIAS is used to bias the sensor within the input range of the ADC. The most common type or sensor application that would use VBIAS is a thermocouple. Although other types of sensors can be used the VBIAS has limited sinking/sourcing capability.
Similarly the burn-out current sources (BOCS) are used for detecting opens/shorts in the sensor path and most commonly used with thermocouples.
I would suggest taking a look at A Basic Guide to Thermocouple Measurements:
http://www.ti.com/lit/an/sbaa274/sbaa274.pdf
Section 2.4 shows an example of VBIAS and BOCS with thermocouple measurements. There are other combinations also shown in the guide.
Best regards,
Bob B