Tool/software:
We are using the EVM430-FR6047 to evaluate a water flow measurement system using transducers clamped on to the outside of a water pipe. We are getting good results but have an issue we don't understand. That issue is the calculation of the spacing (or separation) between 2 transducers on a given pipe. We are using Snell's law, the dimnesions for a given pipe, and geometry to calculate the transducer spacing for V configured transducers. We have discovered that the calculated spacing is a starting point followed by increasing the separation a little bit at a time to maximize the amplitude of the received signal in the ADC Capture tab of the USS software spplication. Once the spacing has been optimized, then we adjust other settings to optimize the performance for making TOF measurements.
From our experince so far, it seems we always have to increase the transducer spacing from the calculated spacing using the process mentioned above. Increaseing the trasnducer separation is needed regardless of the transducer configuration we are using in a particular instance: Z, N, or V. Incidentally, we've never been able to make the W configuration work - there just isn't enough signal.
Can you provide some insight or comments regarding having to increase the transducer separation from the claculated spacing?