Hi
Please give us your opinion for the following usecase?
Our customer is about to measure a distance approximately 1cm to 100cm in the air.
We have TDC1000-TDC7200EVM.
1. Is it possible to realize above application using TDC1000 solution?
2. If it is possible, what kind of transducer can be recommended?
BestRegards
Hi Akeem,
We are looking at a similar application. We are trying to sense the level of liquid in a bottle, so the range is from around 1cm to 20cm, and it would be great to be accurate to +- 1mm.
The TDC1011 IC seems ideal for this, however, our sensor will sit on the top of the bottle, so really we would be measuring the distance up to the surface of the water (where we should have a reflection). Do you still recommend the PGA460 over the TDC1011 for this?
Many thanks,
Pambo
Hi Akeem,
Thank you very much for your very comprehensive reply!
We were looking at a bistatic solution with 300khz transducers. I will follow your advise and investigate further.
Our application is handheld and power efficiency is very important. Do you have any recommendations or a starting point on the driving voltage for this range (1-20cm) with a 300khz transducer?
Many thanks,
Pambo
Hi Pambo,
Given your maximum range is only 20cm, you can minimize your driving voltage to the PGA460's 6V operating requirement to ensure you also minimize the transducer's ringing decay effect to help with your minimum range measurements. You can actually supply the center-tap of transformer with a lower voltage than the PGA460's 6V minimum requirement.
The PGA460 GUI contains an embedded Power Budget Calculator. Assuming you are pinging the system at a 1Hz rate, I recommend that you use a load switch to completely disconnect power between each ultrasonic measurement. The PGA460's minimum preset record length is 4.096ms (or approximately ~70cm) for a single burst-and-listen cycle. Using the UART interface at 115.kBaud will require the PGA460 to be powered for an additional ~2ms to send the command and read the results, and an additional ~3ms to bulk configure at initial power-up. All three steps (1. bulk threshold write, 2. burst/listen command, 3. pull results) yields a total powered time of 9~10ms.
Here is an example of the power budget calculation for the high-frequency configuration recommended for this application:
Hi Akeem,
Thats very useful, and thank you for the sample power budget! That gives us very valuable insight at this stage.
Based on this, I wonder if we could use the TDC1011 instead so that we can operate at a lower driving voltage and reduce the power consumption and ringing decay further? Or would you not advise going much below the 6V mark for a range of up to 20cm?
Many thanks again,
Pambo
Hi Akeem,
I'm inclined to proceed with the PGA460 actually. It might be worth asking your colleague who specialises in the TDC about the driving voltage, just to have an idea of what power savings can be made, however it does seem like the consumption is already acceptable from what you've sent previously!
All the best,
Pambo