Part Number: TDC1000-TDC7200EVM
Hello,
I am designing a low-volume flow sensor for water with the TDC1000-7200EVM and 1MHz transducers (STEMINC SMDM21F1000, https://www.steminc.com/PZT/en/ultrasonic-flow-sensor-1mhz-send-receive). I cannot, no matter the settings, achieve a Delta ToF close to zero. I can see clear START, STOP, and COMP_IN signals on an oscilloscope, but the values output in the GRAPH tab of the GUI are always high (~300 ns) with a significant STDEV. I’m testing with deionized water at zero flow. Additionally, the ToF I measure on the scope varies with the AUTOZERO_PERIOD set in the GUI; shouldn’t that be a function of distance rather than a register setting? Any help regarding these matters would be much appreciated.
Hansen
Hi Hansen,
To address the final question first, if the autozero setting is affecting the TOF, then this means that you are masking the received signal. The receive circuitry is not fully active until the autozero period is complete, so if the autozero setting is long enough, it will prevent the device from seeing the echo signal until it is complete. Please refer to this measurement sequence diagram:
If the autozero is masking your return signals in this way then it will also affect the delta TOF.
Additionally, mismatch between transducers will cause the delta TOF to drift away from zero in a zero flow condition. I am linking two application notes that address this issue and provide some examples of analog interface circuits that can mitigate this problem. One is based mostly on gas flow sensing, but both have pertinent information on this topic.
Regards,
Scott,
I've attached the configuration file I'm using for testing. I am not using an RTD in my setup.
Thank you,
Hansen
TDC1000_Config0 48 TDC1000_Config1 45 TDC1000_Config2 12 TDC1000_Config3 03 TDC1000_Config4 06 TDC1000_TOF-1 A4 TDC1000_TOF-0 26 TDC1000_Error_Flags 00 TDC1000_Timeout 33 TDC1000_Clock_Rate 00 TDC1000_Conts_Trigger 01 TDC7200_Config1 02 TDC7200_Config2 04 TDC7200_Interrupt_Status 00 TDC7200_Interrupt_Mask 07 TDC7200_Coarse_Cntr_Ov_H FF TDC7200_Coarse_Cntr_Ov_L FF TDC7200_Clock_Cntr_Ov_H FF TDC7200_Clock_Cntr_Ov_L FF TDC7200_Clock_Cntr_Stop_Mask_H 00 TDC7200_Clock_Cntr_Stop_Mask_L 00 TRIGGER_UPDATE_FREQ 0000 TDC_AVG/STDEV_NUM_ELEMS 000A SAVE_GRAPH_DATA_TO_FILE 00 FLOW_MODE_SELECT 01 Y-SCALE_CENTER 0032 X-SCALE_RANGE 0032 MEASURE_RTD1 00 SAVE_RESULT_REGR_TO_FILE 00 GRAPH_MULTI_STOPS 00 TDC_SELECT 00 TEMP_RTD_SELECT 00 TEMP_RTD_MODE 01 TEMP_AVG/STDEV_NUM_ELEMS 000A SAVE_TEMP_DATA 00 GRAPH_YMAX_DATA 0000000000000258 GRAPH_YMIN_DATA FFFFFFFFFFFFFD44 TDC1000-HV_DRV_EN1 00 TDC1000-HV_DRV_EN2 00 HV_DRV_EN1_Period 001E HV_DRV_EN2_Period 001E ENABLE_POWER_CYCLE 00 CLK_FREQ_(1-16MHz) 8.000000 CLK_SOURCE_SEL 00 CPU_CLK_FREQ 05 CPU_CLK_EN 00 TDC1000-HV_BST_PWR_EN 00 HV_BST_PWR_EN_Period 0000 FWD2REV_FLOW_DELAY 0000 TDC1000-IMPE_MATCH_EN 00 ENABLE_UART_STREAM 00 ENABLE_MSP430TIMER_TDC 00