Hi, Dale Li:
I read your post ADS131M08: Simultaneous sampling ADC selection for strain measurements.
What about the 300K differential input impedance on ADS131M08? Does it affect the bridge measurement?
Thanks.
-WQ
This thread has been locked.
If you have a related question, please click the "Ask a related question" button in the top right corner. The newly created question will be automatically linked to this question.
Hi, Dale Li:
I read your post ADS131M08: Simultaneous sampling ADC selection for strain measurements.
What about the 300K differential input impedance on ADS131M08? Does it affect the bridge measurement?
Thanks.
-WQ
Hi Wenwei,
Bridge resistance is typically less than 3kΩ, 120Ω, 350Ω, and 1kΩ are the most common values for a bridge. They are small comparing with 330kΩ input impedance of ADS131M08, also a system calibration is usually done, so it will not have any significant impact for the measurement. The following app note can be a good reference for you:
A Basic Guide to Bridge Measurements
Regards,
Dale