Other Parts Discussed in Thread: DDC264
Hi Team,
We would like to ask your help regarding our customer's inquiry below.
My team and I are currently working on the hardware design of a handheld luminometer. For the measurement process we found out that the best solution for the current levels that we will be measuring is the IVC102 with a S1227-66BR Photodiode. On our project concept stage the following circuit was considered and tested:
In order to compensate the S1227-66BR’s dark current and subtract it from the measurement, it was proposed to actually measure this current with the use of a shutter device before and after each measurement cycle.
Recently I came across a Medical Applications Guide from 2007 produced by TI (https://www.mouser.com/catalog/specsheets/tismedicalappsguide.pdf) where a different circuit using two IVC102 is proposed claiming common mode errors are subtracted.
From a theoretical stand point, I can see that the circuit will actually mitigate the Dark Current, but I would like to ask if the requirement of a Shutter device on my previous circuit can be ignored with this improved configuration. Furthermore, I would like to know if I should have some additional considerations to the paired IVC102 of the improved circuit. Should the ICs be from the same batch/bin? Should we test the devices be tested in some way to prevent excessive parameter dispersion? How would it affect?
Additionally, checking the datasheet I see that a carefully selected, low noise power supply should be used for this IC, as with any analog measuring device. In that regard, I would appreciate if you could assess a proper supply from TI’s catalog for this application (handheld device, input voltage +5 V).
Finally, is there a pre-selected Analog to Digital Converter with I2C interface that is recommended for this circuit?
Regards,
Danilo