Part Number: TDC1000-GASEVM
I'm currently conducting an experiment to see how the TOF varies with O2 Concentration.
I have a TDC1000-GASEVM/TDC1000-BSTEVM and am using the TDC1000_7200_EVM Software.
However, the graphical output of the measured TOF is very noisy and fluctuating by about 100ns, when it should be quite constant if the O2 concentration is constant (i'm using a portable oxygen concentrator to provide flow through the sensor)
It also outputs a fluctuating TOF when the sensor isn't connected to any oxygen source.
Hi Scott,
After completing more work on this today, i'm not so worried about the fluctuations, but more concerned about if my TOF is correct for my transducer measurements.
Please find attached my config file and screenshots of graphs at 90% and 20% O2. Additionally, my transducer is approximately 5.5cm, so using the formula TOF = distance between transducers/speed of sound I am not getting the correct values.
TDC1000_Config0 04 TDC1000_Config1 41 TDC1000_Config2 20 TDC1000_Config3 0D TDC1000_Config4 5F TDC1000_TOF-1 40 TDC1000_TOF-0 1E TDC1000_Error_Flags 00 TDC1000_Timeout 23 TDC1000_Clock_Rate 01 TDC1000_Conts_Trigger 00 TDC7200_Config1 82 TDC7200_Config2 40 TDC7200_Interrupt_Status 06 TDC7200_Interrupt_Mask 07 TDC7200_Coarse_Cntr_Ov_H FF TDC7200_Coarse_Cntr_Ov_L FF TDC7200_Clock_Cntr_Ov_H FF TDC7200_Clock_Cntr_Ov_L FF TDC7200_Clock_Cntr_Stop_Mask_H 00 TDC7200_Clock_Cntr_Stop_Mask_L 00 TRIGGER_UPDATE_FREQ 0000 TDC_AVG/STDEV_NUM_ELEMS 00C8 SAVE_GRAPH_DATA_TO_FILE 00 FLOW_MODE_SELECT 00 Y-SCALE_CENTER 0032 X-SCALE_RANGE 0032 MEASURE_RTD1 00 SAVE_RESULT_REGR_TO_FILE 00 GRAPH_MULTI_STOPS 00 TDC_SELECT 00 TEMP_RTD_SELECT 00 TEMP_RTD_MODE 00 TEMP_AVG/STDEV_NUM_ELEMS 000A SAVE_TEMP_DATA 00 GRAPH_YMAX_DATA 00000000000216FC GRAPH_YMIN_DATA 0000000000021698 TDC1000-HV_DRV_EN1 01 TDC1000-HV_DRV_EN2 01 HV_DRV_EN1_Period 012C HV_DRV_EN2_Period 012C ENABLE_POWER_CYCLE 00 CLK_FREQ_(1-16MHz) 2.000000 CLK_SOURCE_SEL 03 CPU_CLK_FREQ 0C CPU_CLK_EN 01 TDC1000-HV_BST_PWR_EN 01 HV_BST_PWR_EN_Period 0000 FWD2REV_FLOW_DELAY 0000 TDC1000-IMPE_MATCH_EN 00 ENABLE_UART_STREAM 00 ENABLE_MSP430TIMER_TDC 00
TDC1000_Config0 04 TDC1000_Config1 41 TDC1000_Config2 20 TDC1000_Config3 0D TDC1000_Config4 5F TDC1000_TOF-1 40 TDC1000_TOF-0 1E TDC1000_Error_Flags 00 TDC1000_Timeout 23 TDC1000_Clock_Rate 01 TDC1000_Conts_Trigger 00 TDC7200_Config1 82 TDC7200_Config2 40 TDC7200_Interrupt_Status 06 TDC7200_Interrupt_Mask 07 TDC7200_Coarse_Cntr_Ov_H FF TDC7200_Coarse_Cntr_Ov_L FF TDC7200_Clock_Cntr_Ov_H FF TDC7200_Clock_Cntr_Ov_L FF TDC7200_Clock_Cntr_Stop_Mask_H 00 TDC7200_Clock_Cntr_Stop_Mask_L 00 TRIGGER_UPDATE_FREQ 0000 TDC_AVG/STDEV_NUM_ELEMS 00C8 SAVE_GRAPH_DATA_TO_FILE 00 FLOW_MODE_SELECT 00 Y-SCALE_CENTER 0032 X-SCALE_RANGE 0032 MEASURE_RTD1 00 SAVE_RESULT_REGR_TO_FILE 00 GRAPH_MULTI_STOPS 00 TDC_SELECT 00 TEMP_RTD_SELECT 00 TEMP_RTD_MODE 00 TEMP_AVG/STDEV_NUM_ELEMS 000A SAVE_TEMP_DATA 00 GRAPH_YMAX_DATA 00000000000216FC GRAPH_YMIN_DATA 0000000000021698 TDC1000-HV_DRV_EN1 01 TDC1000-HV_DRV_EN2 01 HV_DRV_EN1_Period 012C HV_DRV_EN2_Period 012C ENABLE_POWER_CYCLE 00 CLK_FREQ_(1-16MHz) 2.000000 CLK_SOURCE_SEL 03 CPU_CLK_FREQ 0C CPU_CLK_EN 01 TDC1000-HV_BST_PWR_EN 01 HV_BST_PWR_EN_Period 0000 FWD2REV_FLOW_DELAY 0000 TDC1000-IMPE_MATCH_EN 00 ENABLE_UART_STREAM 00 ENABLE_MSP430TIMER_TDC 00
Please find attached my oscilloscope readings.
As you can see, there's a clear sending pulse (green) however the receiver side (yellow) does not have the same pulse in response. I was thinking its because the signal has dissipated as it travels through the transducer.
Additionally, just to clarify i also attached a picture of my transducer. As you can see it has a sender and receiver on each side, so should the TOF still be about 300us?
Thank you.
Kind regards,
Hi Scott,
I tried on both mode 0 and mode 1 (with damping disabled)
However, on mode 1 my TOF is around 390us, and on mode 0, my TOF is around 150us. Mode 0 also has a much lower standard deviation of approximately 20.
Thanks,
Madeleine
Additionally,
Is there a way for the program to output data showing the real time information? (i.e. another column that shows the real time next to the TOF measurement)
And is there a way to edit this code? I'm assuming this interface was created on LabView or something similar.
Thank you.
Madeleine
Hi Scott,
I used mode 2 to attain the TOF (the screenshot of the graph is attached)
The delta TOF fluctuated between around 200 and -200, are these acceptable values?
I've downloaded IAR embedded workbench for MSP430, can the board be connected straight to the PC via the usb for programming or does it require another serial interface?
By real time information, I was wandering if there's a way for the outputted excel file to display the real time in one column. For example if I wanted to know what the value of the TOF was at exactly 60 seconds, I could know what the value of the TOF was 60 seconds into running the test from the outputted excel sheet.
Thank you for your continued support!
Kind regards,
Madeleine
Hi Madeleine,
Is the air/02 mixture flowing during this measurement? The concentration measurements can be very sensitive to turbulence in the measurement chamber.
Yes, you can program the MSP430 using the USB interface on the EVM. You will have to follow the same procedure that is described in the EVM user's guide for updating the firmware on the MSP430.
Each line in the excel sheet is one measurement that is taken at the rate set by the GUI (up to one per 100ms). You can create a new column in the sheet with the first measurement at 0, and each subsequent measurement will be at the time equal to the previous time + your measurement rate selected in the GUI. The GUI itself does not have a way to output this information however.
Regards,
Hi Scott,
Sorry for the late reply.
Yes both the air and O2 gas was flowing through the chamber when taking the measurements.
Would this effect the accuracy of my results a lot?
Kind regards,
Madeleine