Part Number: TDC1000-TDC7200EVM
Hi,
I have a TDC1000-TDC7200EVM development kit to provide a MVP.
My main goal is measure liquid level through the bottom of a metal tank.
The first test with a plastic cup was nice! I'm able to detect a great range of the liquid level. Follows the image:
After these first test, I changed the plastic cup to a metal bottle, 7 cm of diameter and 20 cm of height. I set the TDC1000 register like the suggestion of a "good start" from Matthew Minasi in this thread: https://e2e.ti.com/support/sensor/ultrasonic/f/991/t/467713
My transducer was installed like Matthew Minasi thread.
I don't know why I'm not able to get the same result. I debugged the signal with my oscilloscope and the TDC1000 can't even generate the STOP signal.
The image below is the START and COMP_IN signal.
Can anyone help me to find what am I doing wrong?
Thank you very much!
Hi Bharat,
I will make some tests today and I let you know ASAP.
Thanks for the tip!
Hi Bharat,
I bought a new metal tank with 300 mm of height and 200 mm of diameter, to improve the tests.
I did a lot of tests as you asked, changing the RX COUNT with this configuration, but none of the 7 options changed the result.
The STOP signal didn't happen either.
The image bellow is the COMP_IN (Yellow) with no stop signal (Blue).
After this test, I did some modifications:
Doing this I could to see the "stop signal" with the tank full of water.
I started to empty the tank with a rubber hose in a constant flow. The image bellow is the result.
As you can see, the measurement is not constant like the plastic cup.
In the oscilloscope, I can see a lot of stop signal, but changing fast in the time generating "spikes" in the graphic. If I increase the ECHO_QUAL_THLD to -220mV or higher, I not able to generate the stop signal again.
Thank you!
Regards,
Hello Uilter,
Looks like the return Tx signal is week we can approach this through either of the 2 ways as below:
1. Try to increase the Tx voltage thorugh an external amplifier or provide the TDC1000 with a 5.5V (remember the Jumper JP3 in the EVM connects the VDD & AVDD, please remove this jumper & then connect a 5.5V to the VDD supply pin.
2. Also try a large dimension transducer.
please let me know if you need any further info in this regards, I will be happy to help you.
Hello Bharat,
I'll try both options to get a better result.
ASAP I'll return with the progress.
Thanks.
Hi Bharat,
The option 1 improved a litle the signal, but in high levels the signal have a lot of noise. This "noise" could be generated by a too low ECHO_QUAL_THLD?
The option 2 was the worst, I was able just to read 50% of the level, the rest I get a timeout. I tried this with a 25mm transducer.
Measuring the COMP_IN there is no signal at all. With large transducer I need to change any settings?
Thank you!