Software Tool does not work with ADS131A04EVM

Other Parts Discussed in Thread: ADS131A04EVM, ADS1261EVM

Hi,

i use Windows10 with .Net Framework (Vers. 4.7.2) and i have troubles to go in connection with the hardware.

First i installed DELTASIGMAEVAL-GUI-installer.exe and after that, i installed ADS131A04-dvcpkg.exe. All seems fine and also Generic Bulk Device Driver was installed sucessful.

When i open first Delta-Sigma ADC EVAL Tool it open, but if i connect the hardware -> no connection was found (Image 1)

When i open Delta-Sigma ADC EVAL Tool after i connect the hardware -> hardware connect, but no Addresses or Registernames are shown on screen (Image 2, 3)

Please help!

  • Haimo,


    I think the problem is that the software can't find the installation of the device package. You would see this if you didn't install the device package (which you say you have). However, the other possibility is that the installation of the Delta Sigma Evaluation software was installed by one user, and the package was installed by another; or that the device package wasn't installed with an admin account.

    Regardless, there is a useful post on this problem at the following link:


    The link describes this problem for the ADS1261EVM, but it applies to the ADS131A04EVM as well.

    Read through the post and follow the instructions and you should be able to get the ADS131A04EVM working.


    Joseph Wu