Regarding programing set points for the "USB-DAQ EVM" when calibrating the PGA309.
I have been using the "USB-DAQ EVM" to calibrate the "PGA309" signal conditioning IC for use with a strain bridge with good success. I have used the "USB-DAQ EVM" successfully as a single unit calibrator and as the brain of a Multi-Cal system calibrating up to 24 PGA309's. We have purchased 8 of the "USB-DAQ EVM" boards again with good success. About 6 months ago one of the "USB-DAQ EVM" units started to insert a slight offset to the calibration point. Enough to cause unacceptable calibration errors. This was a unit used only for testing. Not sure to the cause, I replaced the onboard AtoD thinking there was a problem with it. Then I reloaded the software. The offset persisted and the software reload didn't help. To continue using the unit for calibration experiments, the offset was incorporated into the set point to get a better calibration. (The offset was about .093v high at the F.S. or 4.593. So the set point was reduced by .093 or 4.407. it wasn’t perfect but close enough for my testing.) When I used a different "USB-DAQ EVM" the offset went away. I believed this to be a one off and didn't pursue it further.
One month ago a similar offset started occurring with our 24 channel calibrator. Note this unit had previously been working fine for some time. This is a serious problem as it is part of our production line. Note that the Multi-Cal system no longer uses the "USB-DAQ EVM" onboard AtoD as a voltage measurement it uses the HP34401 multimeter. The symptom is still the same.
Has there been any other problems like this? Do you have any ideas how to fix it?
I have attached some images showing an independent meter monitoring the actual set point and the indicated voltage in the software. They are clearly different and I have used many other meters to check this. They all show the same offset. This offset is real and it is programed into the PGA309. Again, if I use a different "USB-DAQ EVM" the problem goes away.
Joseph,
For production PGA309 calibration you should use a digital multi meter. If you look on page 34 of the users guide, it describes how to do this ( see PGA309EVM users guide) . The AtoD on the USB-DAQ EVM are intended for proof-of-concept and not production testing. Furthermore, these AtoD are calibrated when the USB-DAQ EVM are built, so replacing the AtoD will not resolve calibration issues. The multi-cal has a similar feature that allows for use of an external DMM. The software is designed to accept a 34401 DMM. It can work with others, but this will require some driver development. Do you have a 34401 DMM?
I hope this helps. I suppose it is possible that some other issue is causing the offset, but you should be using a DMM for production calibration anyhow.
Best regards,
Art
Please note that we are using the HP 34401 DMM for calibration. I only brought this issue up when we started to have a problem with the calibration system. Note that I stated that I let the problem slide on the development kit. This issue is now manifesting on the calibration system that is using the HP34401. It did not do this at first. Please note the second paragraph in my first message.
thanks
Joseph,
Sorry, I missed the point that you are using the 34401 with the multi-cal. With this in mind it is not likely that you are seeing issues with the actual measured value out of the PGA309. Here are some possibilities:
Suggestions/questions:
I hope that I can help you work through the problem soon.
Best regards, Art Kay